Arran Cameron:
Plastic class 1 devices where the earth conductor of the mains cable is connected to a metal chassis inside, or is used for screening EMI, has to be tested as class 1 rather than class 2 because of the potential for earth leakage current that could trip a RCD.
There was a (at least one) case of a plastic class 1 device in a hospital that would trip a RCD because of a fault resulting in excessive earth leakage current, but it successfully passed the class 2 EST that was being used at the time.
Is that a safety issue? I can, however, see that in hospitals, more attention is absolutely necessary with equipment that may be used in a Group 1 or Group 2 medical location - but I'd expect a greater degree of competence to test and maintain that equipment anyway?
In any case, even testing Class I appliances in that way won't always detect issues that cause RCDs to trip - or even detect all faults.
For example, if you test a washing machine, a faulty component or circuit may not be switched "on" until the door is closed and the interlock connects, and the component is in circuit set by the control board.
gkenyon:Is that a safety issue?