Hello,


We have a Insulation Resistance measurement unit from Hioki - ST5520.

We are using this to measure multiple devices controlled through a relay matrix as shown in the figure. The positive and negative terminals of the ST5520 is multiplexed through the relay matrix to the 64 pins of the connector. There are 7 pcb's each containing 20 relays.

The individual board testing from ST5520 to the connector is giving a open condition result with a high value of insulation resistance of 9.90Gohm @1000V, when there is no load at the connector output, it is left open.

But the when the whole system is wired and tested with the 7 pcb's(all 32 channels), the value fluctuates and starts with a value from 100Mohm and eventually reaches open condition value.

Is there a problem with this process of multiple channel testing? or this is the expected behaviour? the open condition value(9.90Gohm@1000V) is maintained once it reaches this value and further tests result in this same value. 

Once the system is left OFF for a period of a day, then the testing is resumed again, the same behaviour of starting from 100Mohm to open condition value repeats. Is this to be expected or i am understanding/ doing something wrong?

Any help with this is much appreciated. Thanks in advance.
0d3eb1e11f9f4ba9176cae1b22dc1e68-huge-insulation-measurement.png
Parents
  • The reverse effect depends on the mechanism - there may be no stored charge - if it was mostly caused by current flowing to drive off damp then although you put charges through the system to do that, you do not get them back however long you wait, as that is essentially a resistive effect. However, the low resistance will be back in a few hours of days if you do not control the environmental conditions.

    If it is dielectric absorption and recovery (loose but not free charges) then the recovery time is similar to the the charge up time.

    I;d also question how the parts are going to be used after testing - if 100nA really matters, they will pass on the day, but may well fail after a few days in storage.
Reply
  • The reverse effect depends on the mechanism - there may be no stored charge - if it was mostly caused by current flowing to drive off damp then although you put charges through the system to do that, you do not get them back however long you wait, as that is essentially a resistive effect. However, the low resistance will be back in a few hours of days if you do not control the environmental conditions.

    If it is dielectric absorption and recovery (loose but not free charges) then the recovery time is similar to the the charge up time.

    I;d also question how the parts are going to be used after testing - if 100nA really matters, they will pass on the day, but may well fail after a few days in storage.
Children